Groundwater Arsenic Collection Containers and Preservation for ICP-MS Analysis

729 Views

I am working on my master's of water resources science research in hydrogeology and will be collecting samples from wells (agricultural, livestock, and residential) to analyze for TOTAL arsenic.  

I have conducted a thorough review of the literature and consulted two of the lab managers I will be working with, but I still do not have a clear answer to the proper collection procedures.  I will initially be screening the samples with a quick test kit and a colorimeter, then collect samples (60ml should be sufficient) to use for analysis in ICP-MS.  

My question is, is HDPE or polypropylene used to collect the samples?  Since I will only be preserving the samples for less than a week, does it matter which one I use?  Additionally, multiple methods refer to solely nitric acid as a rinse and preservation (ph <2) while others refer to both HCL and nitric acid.  

Will the Cl in the HCL interfere with any ICP-MS analysis to ensure an accurate total arsenic analysis?  I would additionally appreciate any references to materials as a guide.  

Thank you all in advance for your time.

Answers

Comments

Anonymous
I write to introduce our financial services to you. I am a direct mandate to a Financial Institution who is also known as private lender specialized in the Lease and Sales of Bank / Financial Instruments in the form of Bank Guarantees (BG), Standby Letter of Credit (SBLC), Documentary Letter of Credit (DLC), Letter of Credit (LC), Performance Bonds (PB), Medium Term Notes (MTN) et al with funds is purely earned from private and corporate investment portfolios without criminal origin.

Contact : Mr Paul Wung
Email: paulwung.mandate@gmail.com
Skype ID: paulwung.mandate
Isabela Adriana Calin
Hi

'Groundwater Arsenic Collection Containers and Preservation for ICP-MS Analysis' - I can recommend you a new, simple, inexpensive and without side effects - ultrasonic method of Swiss companies. If you want more details, write to me on email: isabela_calin2004@yahoo.com
Abdelsalam Asweisi
Determination of As in biological samples by ICP-MS is difficult for two reasons: the formation of the molecular ion 40Ar35Cl, which overlaps with monoisotopic As at a mass-to-charge ratio (m/z) of 75 (causing spectral interference), and signal enhancement due to organic matrix (nonspectral interference). Available procedures were examined, including the application of different correction procedures using 40Ar37Cl and 16O35Cl molecular-ion formation

Dr Abdelsalam Asweisi
Ademar Cesar Ferreira
I agree with Walter and best way is to select analysis method and sampling together because one can interfere with other. Carefull sampling leaves you to a good conclusion avoiding time wasting because of mistake. Be carefull.
Walter Weinig, PG, PMP
I'd go with whatever the lab manager you're running the samples with says with respect to bottles and preservatives. Issues with interference will depend in part on your target reporting limits - at very low analytical limits interference is usually a bigger problem. Here's a link to a poster presentation I found discussing interference between Cl and As at low levels: http://www.cala.ca/nw_oct11_pt_challenge_moe_poster.pdf.

Sample collection procedures are something to consider carefully. By total arsenic I assume you mean not filtered for dissolved fraction (as opposed to speciating for the valence state). If you are screening with field test kits, I assume you are not going for ultra-low detection limits so conventional sample collection with latex or nitrile gloves etc. should be fine. USEPA has developed procedures for low-level metals sampling, mostly related to aquatic-life standards and especially for mercury. You could look those up if you want to make sure your sampling protocol is appropriate.

Finally, don't forget your quality assurance/quality control (QA/QC) samples. You'll want some duplicates and field blanks to convince yourself what you're measuring is reproducible and real. Also consider equipment rinse blanks for any re-usable equipment that has been decontaminated between samples. For laboratory QA/QC, you'll need some fraction of samples collected at double or triple volume to do matrix spike/matrix spike duplicates. Look up the USEPA Contract Laboratory Program procedures, and their SW-846 analytical protocols for more background.
Jorgen Jonsson
Hi Lauren,

I would be surprised if the choice of bottle material had any impact on your results. I would personally take the cheapest one. You could always do a quick test with 2 bottles of each type, store them with an arsenic solution and analyse the the solution after a week.

Regarding the acid, I would chose either HCl or nitric acid. Again, shouldn't have an impact on the results.

Kind regards,
Jorgen